Why is the OD reading for my duplex lower than what is stated on my Certificate of Analysis?

The extinction coefficients of the single strands are added together to provide the final data on our Certificates of Analysis. When oligonucleotides are annealed; however, these constants are not entirely additive which can cause confusion when calculating the amount of product present in a sample.  Duplex extinction coefficients can be 15-25% lower than expected due to the hyperchromicity of the duplex.  Therefore, we provide a calculation on our duplex Certificates of Analysis to help better determine the actual amount of duplex present in a sample:

(Measured ODs/Total ODs of single strands) x Total Extinction Coefficient =  Duplex Extinction Coefficent